Borhan Kalash

hispeed nxi HPF

October 29, 2013 04:13



Dear frinds the hispeed Nxi is working ok but it has an error come up during daily Cal.(step 16 ).

Note that it is an old error.

warmup HPF(ratio)is out of range:531ch.value--1.260820e -02
What do you think the cause of the a.m. error ?



Regards

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

re: hispeed nxi HPF

October 29, 2013 11:08

This means too much noise in such channel...
Maybe a electronic problem at detector electronics..however, if you cannot see artifacts at image, dont worry.

Log inor Register

to rate and post a comment

(34)
(5)
(4)

Mario Icherenski

re: hispeed nxi HPF

October 30, 2013 09:19

If you make CAM amplifier linearity test you will see that the noising only 531 channel or all channels. Also check the voltage range of both 5V, +12V and -12V DAS power supplies. It is possible that if the X-ray tube labored much, the noise coming from it.

Log inor Register

to rate and post a comment

(183)
(13)
(2)

Borhan Kalash

re: re: hispeed nxi HPF

November 05, 2013 02:32

THANKS
HOW CAN I DO CAM amplifier linearity test?
DO IT NEEDS SERVIES KEY?

ALSO I HAVE NOW ANOTHER ERROR (INTERMETENT)
DAS ERROR
CODE 30-1010-01
ERROR DETECTED BY:CIF
ZERO DAS DATA WAS DETECTED

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

re: re: re: hispeed nxi HPF

November 06, 2013 07:57

Zero data means that the ldetector level drops to no signal, with Xray working. If your channel is too much noisy, could bring this message.

Log inor Register

to rate and post a comment

(183)
(13)
(2)

Borhan Kalash

re: re: re: re: hispeed nxi HPF

November 07, 2013 12:26

thanks
I got the error even without x ray from offline scan
which part do u think the most involved

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

re: re: re: re: re: hispeed nxi HPF

November 07, 2013 03:14

The most probable fail, is the CAM board amplifier. There is a set od op-amp's that amplifies the signal from detector. If you have a extender card, you could search and isolate the failure.

Log inor Register

to rate and post a comment

(183)
(13)
(2)

Borhan Kalash

re: re: re: re: re: re: hispeed nxi HPF

November 08, 2013 05:06

Thank u very much carlos
I'll try to get one.
Is there a test point i have to mesure on cam?

Log inor Register

to rate and post a comment

(39)
(8)

Raul Elizalde

re: re: re: hispeed nxi HPF

November 14, 2013 12:22

For Error Zero DAS data was detected you firs need to know if you get an Image, Normally you can't because the error appear, you need to move a sw in CIF to see the image and then will see the artifact, may be a ring, may be a band, its depend of the damage. but a this time you can know what is the radio o diameter and you can check what channel correspond to the size of the ring and localize the channel and then localize the cam.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Brand new OEM Philips & Siemens Parts at unbeatable prices

Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty! Email P/N for pricing: sales@Relaxationinc.com (P) 813-925-1314, www.relaxationinc.com

stats

Événements de Vente aux Enchères à Venir

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)