Help for Error code 79 (Lorad Affinity series)

Tags: Mammo, HOLOGIC

(11)
(5)

yikwang

Help for Error code 79 (Lorad Affinity series)

July 18, 2014 05:58

Please Help, I found error 79 tubehead_status_error and the system motor for tubehead control board (mirror, filter, lamp and aperture) move in and out in repeatedly. I cannot test each motor with S1 to S4 button.

Log inor Register

to rate and post a comment

(28)
(1)

Larry Underriner

re: Help for Error code 79 (Lorad Affinity series)

July 19, 2014 09:25

Could be a bad Opto Switch not sensing the flag being in position?

Log inor Register

to rate and post a comment

(11)
(5)

yikwang

re: re: Help for Error code 79 (Lorad Affinity series)

July 20, 2014 09:22

Thank, it done, thank i found the screw on aperture detector board's plate loosen and cause the opto switch not sensing.

Log inor Register

to rate and post a comment

(28)
(1)

Larry Underriner

re: re: re: Help for Error code 79 (Lorad Affinity series)

July 21, 2014 08:49

Glad I could help.

Log inor Register

to rate and post a comment