Dear All,
Can we replace copley controls Gradient 262PN with Performance control GA300 without any issue? what will be the cost of Both Gradient ? can we get new performance control GA300 gradient in india ?

Kindly suggest.

1 person liked this
Log inor Register

to rate and post a comment

(23)
(7)
(14)

Giovanni Lombardo

re: Gradient Replacement AIRIS II

April 28, 2015 09:26

Hi,
I can help you with all the solutions for Hitachi, also in the replacement of the two amplifiers ... write to assistenza@gbsmed.it

1 person liked this
Log inor Register

to rate and post a comment

(39)
(8)
(9)

Bruce Smith

re: Gradient Replacement AIRIS II

April 28, 2015 09:27

Even if you have the capability, It is likely that when you make this change you have created a "new systems" that is no longer covered by the governing bodies registrations that the OEM has for the system. In the US it would require a new 510K be submitted in order for the new system to be used in a clinical environment. Check laws and regulations carefully before proceeding.

1 person liked this
Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI Solutions to Meet Your Team’s Unique Needs

From preventive maintenance through comprehensive cryogen services, BMI-PMI is your trusted partner for new/reconditioned sales, installation, and service of both facility-based and mobile MRI solutions. www.brownsmedicalimaging.com / 800-701-9729

stats

Événements de Vente aux Enchères à Venir

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)