GERARDO ALCANTARA

GE SAM ANESTHESIA

May 26, 2011 04:38

I have a GE SAM module that's showing a "service module cell sev"
error message. Wondering if anyone else has seen this and what the
resolution was. Any assistance would be appreciated

Log inor Register

to rate and post a comment

(19)
(2)
(47)

Boyd Campbell

Re: GE SAM ANESTHESIA

May 27, 2011 09:00

This indicates that the Sevo cell in the gas bench is defective but those modules can only display one error at a time and many times we find that there are other issues once you replace the sevo cell. You also need the special software and all the special gasses to cal the unit once the gas bench has been repaired. If you need anymore assistance just let us know and I can put in contact with one of my SAM specialist here.
Boyd Campbell, CBET, CRES