Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

Ge proteus

(28)
(3)

Michael Kakoulli

Ge proteus

November 26, 2015 10:01

Hi everyone,
I have a minor problem with a GE proteus generator after taking an exposure and then removing the cassette from the table bucky the system intermittently plays a tune a series of notes? . Has anyone come across this before . It doesn't seem to affect system operation.

Regards
Mike

1 person liked this
Log inor Register

to rate and post a comment

(20)
(2)

xrayrepairdude

re: Ge proteus

November 27, 2015 08:12

That sounds like the bucky drive motor that has a delay on driving the grid back to home position after the exposure. It may eventually fail and the generator won't expose on that bucky.

Ken

Log inor Register

to rate and post a comment

(28)
(3)

Michael Kakoulli

re: re: Ge proteus

November 27, 2015 10:57

Hi ken,
will check it out.
Mike

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Ge proteus

November 27, 2015 03:46

check the power supplies on the system I/F board in the top of the Jedi cabinet if the the +5 VDC is at or close to 4.85VDC in the base of there is a power supply if you look at the wires coming off the supply you will see a connector with power off reseat this connector(if you have contact cleaner spray or apply to both sides) then retest the supplies.
the intermittent tone can be a reset tone due to these supplies

Log inor Register

to rate and post a comment

(28)
(3)

Michael Kakoulli

re: re: Ge proteus

November 28, 2015 07:18

Thanks will check at next service visit.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Événements de Vente aux Enchères à Venir

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)