Hi
Intermittently facing error.
Axial scan error. Code: 20-1001-26 error detected by: OGP DAS Trigger of one pulse rotation, between G-Pulse 2, is under specification on axial scan.
System warm up normal sometimes complete scan normal.
Regards.

Log inor Register

to rate and post a comment

(88)
(4)

Edurado Madeiros

re: GE ProSpeed A1/F1

September 06, 2017 10:11

Contact me by ermsismed@gmail.com. I can help you

Log inor Register

to rate and post a comment

Oscar Guetta

re: GE ProSpeed A1/F1

September 06, 2017 10:15

There are 2 G-Pulse boards on Gantry, check renewal parts document (2244774) to find them on page no. 91 (1-9). Check cables from G-pulse board to other parts, and clear photo switches. You can swap the G-pulse1 and G-Pulse2 to check.
I think no problem with Axial motor Encoder. Clear the slip ring, brush and check the fuses on brush.

1 person liked this
Log inor Register

to rate and post a comment

(66)
(8)

Haroon Rashid

re: re: GE ProSpeed A1/F1

September 25, 2017 05:02

Thank you acording to your guidance my problem solved.
Regards

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: GE ProSpeed A1/F1

September 06, 2017 10:15

The potential causes will be G-Pulse2, Axial Motor Enconder or interface between this encoder and OGP including the slip ring.

First clean and check your slip ring and brushes. 10 signal slip rings are equipped on the slip ring frame; 7 & 8 are for DAS trigger signal...check and clean them, check for signal brush contact using ring service tool and oscilloscope.

Check interface between Axial encoder and OGP

1 person liked this
Log inor Register

to rate and post a comment

(66)
(8)

Haroon Rashid

re: re: GE ProSpeed A1/F1

September 25, 2017 05:01

Thank you all. I have done slip rings maintenance and cleanig of G-Pulse sensors now system is working smooth.
Regards

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS MX8000 & Brilliance Systems & Spare Parts In Stock Now!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Événements de Vente aux Enchères à Venir

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)