CRAIG WAMBURA

GE LS Plus kV cals

July 03, 2017 08:07

Dear all,

I have just re-installed software on a GE LS but when I try to do a restore state, the MOD can't restore the cal database.

I cannot perform any scans as the system says "No kV stations have been calibrated"

What can I do since I have no other Savestate?

Log inor Register

to rate and post a comment

(42)
(14)

Pawel Kopysc

re: GE LS Plus kV cals

July 04, 2017 06:28

Make a calibration :)

Regards

Pawel

Log inor Register

to rate and post a comment

(183)
(13)
(2)

Borhan Kalash

re: GE LS Plus kV cals

July 06, 2017 09:23

I do not have good experiance
but try to restore just the cal files.
whats up no. 00963954444992

Log inor Register

to rate and post a comment

(9)

Shiju Karakkoottil

re: re: GE LS Plus kV cals

July 10, 2017 09:14

Hi, If you don't have the save state MOD, you need to do the complete calibrations again. What is the error are you getting while loading save state MOD? Send me the image through whatsap at +91 94440 26607 or by email to shiju@tecsila.com