re: GE LightSpeed with Jedi HP60 generator problem
October 06, 2017 02:26
code 80 points to a hardware error.
80-1422 points to overpressure of tube oil or broken fuse.
1. check fuse on IF board
2. check oil pump
3. check communication between CT IF Board and inverter
4. check cabling
5. check for power IF board failure
6. check tube for oil overpressure
Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty!
Email P/N for pricing: sales@Relaxationinc.com
(P) 813-925-1314, www.relaxationinc.com
Manuel Brugat
GE LightSpeed with Jedi HP60 generator problem
October 04, 2017 09:21
I have a problem with a GE LightSpeed CT scanner.
This system has the Jedi HP60 generator.
the first error is 260118367
Host: On-Board Rotating Processor
An unexpected JEDI State occurred
Jedi Not Ready.
The second error is: 260118260
Host: Jedi
Overpress or blown fuse Error 80-1422
Can someone tell me something sbout this?
Sincerely:
Manuel Brugat/Image Instrumentation, Inc.
to rate and post a comment
VENGE
re: GE LightSpeed with Jedi HP60 generator problem
October 05, 2017 09:04
80-1422h- H4 Power IF com. error
to rate and post a comment
Manuel Brugat
re: re: GE LightSpeed with Jedi HP60 generator problem
October 06, 2017 10:02
Thank you for your answer. I understand there is a problem with the IF board may be. But what is causing it?
to rate and post a comment
CRAIG WAMBURA
re: GE LightSpeed with Jedi HP60 generator problem
October 05, 2017 09:05
Mybe you need to reload the JEDI software
to rate and post a comment
Manuel Brugat
re: re: GE LightSpeed with Jedi HP60 generator problem
October 06, 2017 10:02
Thank you for your answer. How do i load the JEDI software?
to rate and post a comment
CRAIG WAMBURA
re: GE LightSpeed with Jedi HP60 generator problem
October 06, 2017 02:26
code 80 points to a hardware error.
80-1422 points to overpressure of tube oil or broken fuse.
1. check fuse on IF board
2. check oil pump
3. check communication between CT IF Board and inverter
4. check cabling
5. check for power IF board failure
6. check tube for oil overpressure
to rate and post a comment
Douglas Murillo
re: re: GE LightSpeed with Jedi HP60 generator problem
October 11, 2019 10:56
how can i check tube for oil overpressure?
to rate and post a comment
Gabriel Cool
re: re: re: GE LightSpeed with Jedi HP60 generator problem
October 14, 2019 10:17
Check logs
to rate and post a comment