Elias Valdes
GE LightSpeed Ultra initializing problem
September 02, 2010 01:05
Hello:
I have a problem with a GE LightSpeed Ultra recently installed. When starting the system the computer stays at "OC Initializing. Please wait" and it stays there. After a few minutes the computer shuts down by itself. Apparently it loads the Applications, menus and everything seems to be there. After performing a Reconfiguration without regenerating the Data Base I can even see images,but same case, after a few minutes the computer shuts down. I have ran the Octane Diagnostics with the IRIX CD and all tests passed, hinv identifies disks and I ran also the ICE tests, first time ICE tests were run they all passed, second time the data marked an Error. So far I have tried to avoid software reload or Reconfiguration regenerating the Data Base. Any sugestions or tips will be appreciated. Thanks
Best regards,
Elias
to rate and post a comment
Artem Potiy
Re: GE LightSpeed Ultra initializing problem
September 02, 2010 01:46
Dear Elias Valdes
Maybe you have bad sector on HDD. So you need to connect this HDD to another computer and test by special program. I usually use HDD regenerator- this program can repaired bad sectors.
I think this can help
Regards
Artem
to rate and post a comment
Josip Vujnovic
Re: GE LightSpeed Ultra initializing problem
September 15, 2010 09:51
Do you have communication with Gantry?
to rate and post a comment
Elias Valdes
Re: GE LightSpeed Ultra initializing problem
September 16, 2010 02:24
Yes, Hardware initialization is completed, I can even move the table, rotate Gantry for centering lights, tilt. No scans can be performed as I don't have access to anything having to do with scanning in the computer. thanks,
Best Regards,
Elias
to rate and post a comment
Josip Vujnovic
Re: GE LightSpeed Ultra initializing problem
September 16, 2010 03:56
Have you tried open DARC with telnet?
I had the same problem, with the same symptoms as you when I was a failure of DARC!
to rate and post a comment
Johnny Barreto
re: Re: GE LightSpeed Ultra initializing problem
August 27, 2012 03:12
How did you fix this problem? i'm having same issue by now
to rate and post a comment
Elias Valdes
re: re: Re: GE LightSpeed Ultra initializing problem
August 27, 2012 08:31
I replaced IG Board. Perform IG Diagnostics to know if it is good or faulty.
to rate and post a comment
Elias Valdes
re: re: re: Re: GE LightSpeed Ultra initializing problem
August 28, 2012 03:18
Don't forget to check also the SGI Octane or Octane2 compuyer
to rate and post a comment
Johnny Barreto
re: re: re: re: Re: GE LightSpeed Ultra initializing problem
August 28, 2012 03:45
Thank you, I made LFC twice, it seems to be fixed, but I'm having an artifact problem with this system, may I email you about it? Mine is johnnyb@tigo.com.py
to rate and post a comment
Jose Morillo
re: Re: GE LightSpeed Ultra initializing problem
November 13, 2023 10:54
I have the same case but I have not communication with the gantry. No ping with STC or ETC. I required a new Scan Disk (the original one is damaged) and a new transceiver because I do have not a valid way to test it trusty. No laser, no cough move, no changes in gantry leds during initialization
to rate and post a comment