GE HISPEED DX/i error

March 24, 2008 10:24

We have just installed a hispeed dx/i, booting is ok but gantry does not respond. Error was oc command has no response from gantry. We checked all cabling and connections and they are ok, emergency stops and emergency reset doesn't work. RMT CONT ASSY was also replaced but still problem was not resolved. Contactors K2, K3 at PDU are not energized upon start up. Could you tell us the power up sequence? Any suggestions? tnx.

Log inor Register

to rate and post a comment

Ralph Childs

Re: GE HISPEED DX/i error

March 25, 2008 12:51

Francis,

You need to check the cable going to J2 in the PDU very closely. The K2/K3/K5 relays are all related only to Gantry Table power. Make sure that your PDU was not originally wired for an external E-stop as well.

Scooter
scanner1@eim.ae

Log inor Register

to rate and post a comment

(30)
(1)

Francis Lienedes Ayson

Re: GE HISPEED DX/i error

March 25, 2008 07:30

problem resolved...=)