please help....

we have a ct/i ct scan with an octane system with irix 6.5 operating system. this is with a 3.5MHu tube, a G5 collimator and E-Das filter boards.

we are encountering a problem that during fast cal and detailed calibration it will pop up an error of "fastcal calibration error, please exit and retry" or "detailed calibration error". Tube warm up is ok it will finish the process... When we try to scan it will shows a series of ring artifacts...

To solve the problem... we first get the sbc and scan disk in the console for a known good machine but it did not solve the problem.. we also try parts from a known good machine like the octane system, ig board, vme, bit3 and fep boards but the problem is still the same, it cannot do the fastcal.. the next thing we do it replace all the filter boards which include the adc boards and all the boards in cage 1 and 2. this boards are known good parts, but when we try it the problem still the same... then we replace the collimator and align it but the problem still there it cannot do fatscal and detailed calibration..

can someone help us where the problem is coming from? most of error it shows that the primary receiver is not in service or the event router is out of service..

please give us some advice..thanks in advance

1 person liked this
Log inor Register

to rate and post a comment

(14)
(3)

Ignacio Garcia

re: GE CT/i CT SCAN Problem

October 19, 2013 08:27

TRY CHECK YOUR SLIP RINGS MUST BE CLEAN, CHECK BRUSHES,CHECK THE FIBER OPTIC FROM THE GANTRY TO THE OC.
ON THE ERROR LOG, WHAT ERROR NUMBER DID YOU GET WHEN YOU TRY TO MAKE FASTCAL?

Log inor Register

to rate and post a comment

(46)
(4)

marivic

re: re: GE CT/i CT SCAN Problem

October 20, 2013 09:49

we alrerady check the slip ring and its brushes and the fiber optic.it seems ok... the errors that i get is..Error: 200306013- received cal done with status CAL_STATUS_RETRY. ERROR: 200306018- cal_composite command out of sync.command expected=execute cal. received=idle cal_composite..ERROR:200001320-failure sendingevent3110 using event router.the event router received an event that had no primary receiver register. ERROR: 200001301-error trace back

Log inor Register

to rate and post a comment

(14)
(3)

Ignacio Garcia

re: re: re: GE CT/i CT SCAN Problem

October 21, 2013 08:00

give me an email address to send you more info regarding...

Log inor Register

to rate and post a comment

(147)
(22)

Wang Zhixuan

re: GE CT/i CT SCAN Problem

October 19, 2013 08:28

I think you should first reconstruct image use good raw data,if the image is good, you should check the DAS by data.

Best regards.

Mr.Wang zhixuan
email:362689810@qq.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

GE, Siemens, Philips, & Toshiba/Canon imaging parts, training & support!

DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com

DOTmed Certified stats

Événements de Vente aux Enchères à Venir

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)