GE Brightspeed table elevation problem error 260134706
May 04, 2016 04:28
Hi friends! I have a problem with elevation table on Ge ct Brightspeed. When I trying it up now any motion. I have error 260134706: check valve, inverter or drive motion. I change hydraulic elevation pump today, but it not help - the table does not rise up. I was able to pick it up after just 10 attempts from the lowest position. inverter constantly went into the restart and gave error for over current bring: oc1, oc2, oc3 in lower position. I think the problem in it. Table work without any problems in horizontal motion. Can any help me? Thanks!
Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty!
Email P/N for pricing: sales@Relaxationinc.com
(P) 813-925-1314, www.relaxationinc.com
Aleksander Proof
GE Brightspeed table elevation problem error 260134706
May 04, 2016 04:28
Hi friends!
I have a problem with elevation table on Ge ct Brightspeed. When I trying it up now any motion.
I have error 260134706: check valve, inverter or drive motion.
I change hydraulic elevation pump today, but it not help - the table does not rise up.
I was able to pick it up after just 10 attempts from the lowest position. inverter constantly went into the restart and gave error for over current bring: oc1, oc2, oc3 in lower position. I think the problem in it.
Table work without any problems in horizontal motion.
Can any help me?
Thanks!
to rate and post a comment
Pepe Garcia
re: GE Brightspeed table elevation problem error 260134706
May 05, 2016 11:29
help yourself with the table processor disables the Securities and purge lines
to rate and post a comment
Aleksander Proof
re: re: GE Brightspeed table elevation problem error 260134706
May 08, 2016 06:04
Thanks for help!
I find problem in drive line cable
to rate and post a comment