GE Brightspeed table elevation problem error 260134706

Tags: CT, GE

(157)
(29)

Aleksander Proof

GE Brightspeed table elevation problem error 260134706

May 04, 2016 04:28

Hi friends!
I have a problem with elevation table on Ge ct Brightspeed. When I trying it up now any motion.
I have error 260134706: check valve, inverter or drive motion.
I change hydraulic elevation pump today, but it not help - the table does not rise up.
I was able to pick it up after just 10 attempts from the lowest position. inverter constantly went into the restart and gave error for over current bring: oc1, oc2, oc3 in lower position. I think the problem in it.
Table work without any problems in horizontal motion.
Can any help me?
Thanks!

Log inor Register

to rate and post a comment

(205)
(18)
(2)

Pepe Garcia

re: GE Brightspeed table elevation problem error 260134706

May 05, 2016 11:29

help yourself with the table processor disables the Securities and purge lines

Log inor Register

to rate and post a comment

(157)
(29)

Aleksander Proof

re: re: GE Brightspeed table elevation problem error 260134706

May 08, 2016 06:04

Thanks for help!
I find problem in drive line cable

Log inor Register

to rate and post a comment