Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

G E MVP Generator Error codes1301&1321

(26)
(11)
(3)

Daniel Alandar

G E MVP Generator Error codes1301&1321

August 16, 2010 10:43

I have been working on a MVP generator which displays error code 1301 when we select large focal spot and 1321 when we select small focal spot. A local GE rep told me that that indicates a bad tube. I did install a known good tube and still have the same codes. I have do not have an error code chart. Any help would be appreciated

Log inor Register

to rate and post a comment

(84)
(11)
(6)

Galen Hiveley

Re: G E MVP Generator Error codes1301&1321

August 20, 2010 08:50

To have a problem with both large and small filaments, I don't think you have ann x-ray tube problem.

Try checking the hi-tension cable ends. The pins in the end of the cables should be somewhat spread apart to make a good connection. Check the cables ends at both the x-ray tube and at the hi-tension transformer recepticles.