Fujifilm PRIMA CPU37A connection problem after installation

Tags: PACS / DR / CR / Digital, FUJIFILM

(3)

Alessandro Clarizia

Fujifilm PRIMA CPU37A connection problem after installation

June 07, 2022 10:22

Hi all, I have installed a new CPU37A in CR Prima reader, but as soon as I tried to connect with the RU pc, the LAN connection is skipped and the reader is no longer pingable, it has lost the network address.
Is there a possibility of card recovery?
Thanks in advance, it's really urgent!

Log inor Register

to rate and post a comment

(18)
(3)

Amir Jemal

re: Fujifilm PRIMA CPU37A connection problem after installation

June 08, 2022 10:01

Hi Alessandro
- you can reset The IP address to Default IP by using DIP switch on CPU board. S7 - DIP switch.
Please follow the service manual > RU IP address reset page - for more clarification.

Feel free if you have any question.
Good luck

2 people liked this
Log inor Register

to rate and post a comment

(3)

Alessandro Clarizia

re: re: Fujifilm PRIMA CPU37A connection problem after installation

June 09, 2022 11:39

Hi Amire, thank you for your reply
Yes, we actually tried the DIP switch but to no avail.
It would seem that there is a technical bulletin that says that from version 3.5, there are compatibility problems and therefore the CPU card resets and loses the IP.
The reader seems to me to have version 2.1 or 2.5, so it appears in the RU PC.
In fact, the lights on the part of the LAN connector have gone out.
I believe a firmware recovery procedure is required.

Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: re: re: Fujifilm PRIMA CPU37A connection problem after installation

October 26, 2023 11:36

We have installed a new cpu (original from the factory) and the ru shows error 11200 (Service manual error description:" Image signal monitor value (simulated current) error; [During bootup]When diagnosing the simulated data transfer, the A/D conversion data for simulated image were out of range."); The cpu is a newer version than the one the device had, if we exchange it with an old version of another device, the error does not appear. At the same time, if we install this new CPU in a functional device, it generates the aforementioned error, so we conclude that there is an update that generates the incompatibility. The supplier has not given us a response so we are looking for a solution here.
Additionally, if we cancel the error and proceed to scan a chassis, the resulting image is aberrant.
Have you had a similar situation? Solution? Thank you

Log inor Register

to rate and post a comment

(3)

Alessandro Clarizia

re: re: Fujifilm PRIMA CPU37A connection problem after installation

June 09, 2022 11:39

Ah, I forgot, the reader that first performed the boot, then carried out the control of the lamp, motors, etc., now remains stationary

Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: re: Fujifilm PRIMA CPU37A connection problem after installation

October 26, 2023 11:36

We have installed a new cpu (original from the factory) and the ru shows error 11200 (Service manual error description:" Image signal monitor value (simulated current) error; [During bootup]When diagnosing the simulated data transfer, the A/D conversion data for simulated image were out of range."); The cpu is a newer version than the one the device had, if we exchange it with an old version of another device, the error does not appear. At the same time, if we install this new CPU in a functional device, it generates the aforementioned error, so we conclude that there is an update that generates the incompatibility. The supplier has not given us a response so we are looking for a solution here.
Additionally, if we cancel the error and proceed to scan a chassis, the resulting image is aberrant.
Have you had a similar situation? Solution? Thank you

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Source for Sony Medical Displays, Printers & More!

Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.

stats

Événements de Vente aux Enchères à Venir

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)