Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

FCRGO A61 error

(42)
(7)

xtekvet

FCRGO A61 error

November 06, 2018 10:17

FcrGo has an A61 error message. Can still take images. Manual does not give reason for error or how to fix it. Any suggestions?

Log inor Register

to rate and post a comment

(82)
(7)
(6)

Alfred M. Brown

re: FCRGO A61 error

November 07, 2018 08:19

2.15.1 Re-exposure processing (A61-A63)If error E35 (Insufficient tube voltage), E36 (Tube over voltage), or E37 (Discharge detected) is detected, re-exposure processing will be performed up to two times. In re-exposure processing, the values of kV and mAs are the same as normal exposure processing (see the following figure).

Log inor Register

to rate and post a comment

(82)
(7)
(6)

Alfred M. Brown

re: FCRGO A61 error

November 07, 2018 08:19

E35
Insufficient tube voltage{MD:2.15.4}
No
- Make sure that CNKV is attached to the CPU board.- Make sure that there is no disconnection or ground fault in the tube voltage detection circuit.- Check the battery voltage during radiography.
E36
Tube over voltage{MD:2.15.3}
No
CPU board3E1-7,3E2-7: L
- Make sure that there is high-voltage discharge at the tube head. -> Check the waveforms between KV+ and KV+G (GND) and between KV- and KV-G (GND) on the CPU board.- Check whether the adjustment value of E36 is not changed to less value.
E37
Discharge detected{MD:2.15.7}
No
CPU board2D3-7: L
- Make sure that there is high-voltage discharge at the tube head.- Make sure that the main transformer is not layer shorted.- Make sure that +kV and -kV are well balanced.


Log inor Register

to rate and post a comment