Dear sir,
An EUB-450 ultrasonic by Hitachi display white block.Who can give us some advice kindly or support some document?
Thank first!
htdocument@126.com
Hi,
First confirm your monitor. You can give the video output from machine to another monitor and test.
regards,
Shiju
Cura Medical Equipment Pvt Ltd.
shiju@cura.co.in
once you have verified your monitor is not the problem. if the problem is in the normal imaging scan area it is one of the DSC boards, I believe one of 2 for the EUB 450, if the problem is outside the imaging scan area look at the Video I/O Board or CPU. Hope it helps.
Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty!
Email P/N for pricing: sales@Relaxationinc.com
(P) 813-925-1314, www.relaxationinc.com
yi huang
EUB-450
April 14, 2006 01:55
Dear sir,
An EUB-450 ultrasonic by Hitachi display white block.Who can give us some advice kindly or support some document?
Thank first!
htdocument@126.com
to rate and post a comment
Vijai shankar raja
Re: EUB-450
April 18, 2006 07:20
Hi,
First confirm your monitor. You can give the video output from machine to another monitor and test.
regards,
Shiju
Cura Medical Equipment Pvt Ltd.
shiju@cura.co.in
to rate and post a comment
Ray Coombs
Re: EUB-450
April 19, 2006 02:09
once you have verified your monitor is not the problem. if the problem is in the normal imaging scan area it is one of the DSC boards, I believe one of 2 for the EUB 450, if the problem is outside the imaging scan area look at the Video I/O Board or CPU. Hope it helps.
to rate and post a comment
Carlos Duran
Re: EUB-450
April 24, 2006 05:46
hello:
you has a electronic board problem, located on memory image chips, and/or buffers.
you must use oscilloscope to check the data/address lines on the respective boards, to locate the fault.
regards.
to rate and post a comment