Error in Philips MicroDoseMammo

May 08, 2020 11:01

Dear Member,
I need help to diagnose a fault occuring in our Mammogram Make:Philips Model: MDM
The system is trying to perform daily quality test but the test fails and after that the user in not able to perform any kind of scanning.

1 person liked this
Log inor Register

to rate and post a comment

(82)
(15)

Herrmann

re: Error in Philips MicroDoseMammo

May 11, 2020 10:43

Hi Peter,

Have you do all the geometry test in the service tool ?
Then when the aliment or the fild limiter are out of range,
then you can have this.

Kind regards
Willi

Log inor Register

to rate and post a comment

(3)
(1)

Peter Socher

re: re: Error in Philips MicroDoseMammo

May 11, 2020 03:27

We have tried the full calibration after login
But it is getting failed

Log inor Register

to rate and post a comment

(82)
(15)

Herrmann

re: re: re: Error in Philips MicroDoseMammo

May 12, 2020 10:01

Hi Peter,

ok, and then the aliment Check on the Service Tool ?
What is the result ?

Kind regards
Willi