Mr Kumar:
Follow this steps:
1.- There where transiets on the line supply, inintialization took place, check this: push OFF and then ON again.
2.- Switch OFF senographe and place the 400PL7 board on extender, check: optocoupler B2, the voltage measured be 0.09V +- 50 mv, Led D4 should be lit and +5v of Lambda powewr supply should be present on 400PL7, if required replace optocouple B2.
God luck
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
sunil Kumar
Error E190 GE 600T
October 24, 2005 02:56
Any details on GE600T mammo error 190.manual does not show this error.
to rate and post a comment
Manuel Casanova
Re: Error E190 GE 600T
October 24, 2005 08:51
Mr Kumar:
Follow this steps:
1.- There where transiets on the line supply, inintialization took place, check this: push OFF and then ON again.
2.- Switch OFF senographe and place the 400PL7 board on extender, check: optocoupler B2, the voltage measured be 0.09V +- 50 mv, Led D4 should be lit and +5v of Lambda powewr supply should be present on 400PL7, if required replace optocouple B2.
God luck
to rate and post a comment