Your Email:
Kenneth Calhoun
ERROR 253 OEC 9600
March 24, 2012 05:05
does anyone know what to do on a OEC9600 with a bootup error of 253
to rate and post a comment
Juan de la Cruz Martinez
Re: ERROR 253 OEC 9600
March 24, 2012 07:46
This message is associated with a loss of serial communication.
March 25, 2012 12:30
Is this serial communication between the control panel board and the aux board??
thai tan
you should check contact on boards, cables. for me, you reassemble the most boards in the system to maintenance. good luck. Thai
Lenny Place
March 25, 2012 11:07
this is most likely the interconnect cable this error means there is a communication loss to the technique processor board.
Matt Du Vall
March 25, 2012 11:08
This out of the manual *Error 253 occurs when the system does not boot fully and displays the "system error #253" message. This message is associated with a loss of serial communication. Do you want a copy of the manual? reply to: cbet_us@yahoo.com
for the service manual, google search: oec 9600 service manual it will be in a PDF
April 12, 2012 10:37
the problem was a bad connection on the serial cable between the panel control board and Aux board.Their is a FSB# 94-020 on this problem.
April 13, 2012 11:47
Cool.
50+ Stretchers in stock for immediate rentals. Contact us for a today for a quote.
Kenneth Calhoun
ERROR 253 OEC 9600
March 24, 2012 05:05
does anyone know what to do on a OEC9600 with a bootup error of 253
to rate and post a comment
Juan de la Cruz Martinez
Re: ERROR 253 OEC 9600
March 24, 2012 07:46
This message is associated with a loss of serial communication.
to rate and post a comment
Kenneth Calhoun
Re: ERROR 253 OEC 9600
March 25, 2012 12:30
Is this serial communication between the control panel board and the aux board??
to rate and post a comment
thai tan
Re: ERROR 253 OEC 9600
March 24, 2012 07:46
you should check contact on boards, cables. for me, you reassemble the most boards in the system to maintenance. good luck.
Thai
to rate and post a comment
Lenny Place
Re: ERROR 253 OEC 9600
March 25, 2012 11:07
this is most likely the interconnect cable this error means there is a communication loss to the technique processor board.
to rate and post a comment
Matt Du Vall
Re: ERROR 253 OEC 9600
March 25, 2012 11:08
This out of the manual
*Error 253 occurs when the system does not boot fully and displays the "system error
#253" message. This message is associated with a loss of serial communication.
Do you want a copy of the manual? reply to:
cbet_us@yahoo.com
to rate and post a comment
Matt Du Vall
Re: ERROR 253 OEC 9600
March 25, 2012 11:08
for the service manual,
google search: oec 9600 service manual it will be in a PDF
to rate and post a comment
Kenneth Calhoun
Re: ERROR 253 OEC 9600
April 12, 2012 10:37
the problem was a bad connection on the serial cable between the panel control board and Aux board.Their is a FSB# 94-020 on this problem.
to rate and post a comment
Matt Du Vall
Re: ERROR 253 OEC 9600
April 13, 2012 11:47
Cool.
to rate and post a comment