I HAVE A PROBLEM IN SYTEC 2000i, WHEN MAKE A WARM UP PROCEDURES, THE ERROR CODE 101-10 APPEARS INMEDIATLY, I NEED THE ADVANCE SERVICE MANUAL, OR YOUR OPINION ABOUT THIS.
Error 101-10 is an azimuth rotation error. Most probably your azimuth pot has either slipped or is defective. You will also need to scope the encoder output if the pot is okay.
We work with hospitals providing reliable ultrasound systems, probes & parts. By offering sales, service, rentals & repairs, we help our clients access high quality imaging systems while keeping costs low & support simple. Please call 760-834-4777!
Manuel Casanova
ERROR 101 -10 IN SYTEC 2000i
March 01, 2008 06:37
HI ALL:
I HAVE A PROBLEM IN SYTEC 2000i, WHEN MAKE A WARM UP PROCEDURES, THE ERROR CODE 101-10 APPEARS INMEDIATLY, I NEED THE ADVANCE SERVICE MANUAL, OR YOUR OPINION ABOUT THIS.
MANUEL
to rate and post a comment
Ralph Childs
Re: ERROR 101 -10 IN SYTEC 2000i
March 01, 2008 11:03
Manuel,
Error 101-10 is an azimuth rotation error. Most probably your azimuth pot has either slipped or is defective. You will also need to scope the encoder output if the pot is okay.
Scooter
scanner1@eim.ae
to rate and post a comment
Tayfun Canli
Re: ERROR 101 -10 IN SYTEC 2000i
March 06, 2008 10:11
Hi Manuel
Please perform Po,Pg adjustment. Refer to function check and adjustment book which must be located with system.
to rate and post a comment