AKMAL TURAEV

DMR+ Error E51-Lateral Colimator Failure 158/019

May 23, 2014 07:32

Hi all i have problem with dmr+ E51 158/019.How can i solve this issue?
Thanks in advance

Log inor Register

to rate and post a comment

(27)
(3)
(5)

James McGinty

re: DMR+ Error E51-Lateral Colimator Failure 158/019

May 23, 2014 06:40

1)make sure the cover is not stopping motion of the blades
2) recycle system, does motor drive ?
3) if motor drives and blades move, check end of travel sensor

let me know if you need additional help..

askeng@digitecmedical.com

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: DMR+ Error E51-Lateral Colimator Failure 158/019

May 23, 2014 06:40

Lateral collimator: slipping during collimator movement *** 158/019

158/019: check for play between collimator lateral motor axis and pulley (screw not
tightened). Possible problem with 800PL1 (Gantry power board).

Log inor Register

to rate and post a comment

(71)
(13)

gerardo millena

re: re: DMR+ Error E51-Lateral Colimator Failure 158/019

May 26, 2014 09:56

At power-up it runs a test on all motors (collimator, compression and tube). The algorithm for the collimator blades test is the stepper motor makes the blade move so it reaches the optic switch then (not sure if maximum opening or closes)it goes back to the default position. If the motor, optic and mechanical are OK and still gives this error, worst case would be a gantry power board. Many years back to confirm gantry board problem is we interchange the motor and optic connectors in the head (lateral and longitudinal blades). Common problem with this gentry power board is that there is a Resistor Pack surface-mount at the trace side that opens (solder related) due to heat generated by the compression resistors.

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: DMR+ Error E51-Lateral Colimator Failure 158/019

May 24, 2014 09:45

Few days ago i've got same error, i tightened screw of pulley with allen key.

Log inor Register

to rate and post a comment

(17)
(2)

alen kadic

re: re: DMR+ Error E51-Lateral Colimator Failure 158/019

January 31, 2020 10:24

This topic is old but still to ask, I have some error, please can you tell me where are this screw, I can't find it yesterday