When I try to make a exposure on a GE CT/e an error comes up
Host:IRIS Proc: tgp_out Error: 186518
File:: tgp_errors.c Line:312
DAS Error
Code: 30-1010-01
Error detected by: CIF
Zero DAS data was detected
Lightly used system available direct from the end user customer. MONARCH QUEST
Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.
CTe Zero DAS data was detected
December 14, 2012 11:36
When I try to make a exposure on a GE CT/e an error comes up
Host:IRIS Proc: tgp_out Error: 186518
File:: tgp_errors.c Line:312
DAS Error
Code: 30-1010-01
Error detected by: CIF
Zero DAS data was detected
any idea about?
Thank you
to rate and post a comment
Gustavo Heredia
re: CTe Zero DAS data was detected
October 15, 2013 11:46
Dear.'m Having the same error with a Hispeed DXi, you can tell me how this problem fixed it.
thank you very much
to rate and post a comment
Ronald
re: re: CTe Zero DAS data was detected
October 16, 2013 04:41
Friends, clean the "slip ring" and The "carbon brush", mostly are ok.
to rate and post a comment