Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

CT/e single slice

Posted by an Anonymous User

CT/e single slice

December 16, 2013 04:25

Hi every one

Regarding CT/e CT Scanner- GE Make, what is the roll of the front panel board during OS booting ?
The troubleshooting says that OS booting failure may be caused by :
Host PC Power feedback CN4 to front panel fails .
Please advice



Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: CT/e single slice

December 16, 2013 07:07

If front panel has problem, it will show the error message. You can take off the front panel, make a short connnection from the cable, the unit can run without the front panel.

Log inor Register

to rate and post a comment

(67)
(7)

Galal

re: re: CT/e single slice

December 17, 2013 08:27

Thank you for reply.
The connection to panel has 4 lines , please tell me which lines to be short-connected .

Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: re: re: CT/e single slice

December 17, 2013 01:26

There is a big cable which connect the Gantry and FRONT PANEL,take this one out, and put into the service board in upper coner. Where are you located? my Email:lindong123@hotmail.com

Log inor Register

to rate and post a comment

(57)
(3)
(6)

R.K.S singh

re: CT/e single slice

December 17, 2013 08:28

please email me at rohitsingh4u@hotmail.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

GE, Siemens, Philips, & Toshiba/Canon imaging parts, training & support!

DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com

DOTmed Certified stats

Événements de Vente aux Enchères à Venir

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)