Hi every one,
I have problem with vertical movement of the table, it does not move up or down. Only when I press service key in the left side of the gantry the table may move up and down. The gantry can't be tilted or rotated. I appreciate If any body can help me with this or send me the service manual.
Hi Dina!
Check under the table below the integrity of the parts: a rubber ring at the beginning of the table on the motor, and a rope of a vertical potentiometer at the end of the table in front of the gantry
I forget to tell that this problem happened when I switch to service key of the gantry and then use the up & down switches at the bottom of the table. May be some thing happened to one of the limit switches. My question is how many limit switches do I have? Because I recognize 6, and not sure if there is any more?
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
Dina Alnahdi
Brilliance 16
April 08, 2019 10:25
Hi every one,
I have problem with vertical movement of the table, it does not move up or down. Only when I press service key in the left side of the gantry the table may move up and down. The gantry can't be tilted or rotated. I appreciate If any body can help me with this or send me the service manual.
to rate and post a comment
Joseph Jenkins
re: Brilliance 16
April 09, 2019 10:20
email me at iiltd@hotmail.com
to rate and post a comment
Aleksander Proof
re: Brilliance 16
April 09, 2019 10:20
Hi Dina!
Check under the table below the integrity of the parts: a rubber ring at the beginning of the table on the motor, and a rope of a vertical potentiometer at the end of the table in front of the gantry
to rate and post a comment
raj venugopal
re: Brilliance 16
April 09, 2019 10:20
check the ACS unit
to rate and post a comment
raj venugopal
re: re: Brilliance 16
April 10, 2019 08:57
hi send the 7segment display of ACS unit in placed in table top than service mode switches on manuvel up and down is working....?
to rate and post a comment
Indarpersad Singh
re: Brilliance 16
April 09, 2019 10:21
I am not an expert, but you can check Patient Table Board D702. Controls Vertical and also interface between PHS and Gantry.
to rate and post a comment
raj venugopal
re: re: Brilliance 16
April 10, 2019 08:57
you sent the 7 segment led display of ACS control in table
to rate and post a comment
Dina Alnahdi
re: re: re: Brilliance 16
April 10, 2019 04:57
7 segment shows number "3"
to rate and post a comment
raj venugopal
re: re: re: re: Brilliance 16
April 11, 2019 11:52
sir send the error detail my mail id rajvenugopal@greenmedichealthcare.com
send the your mobile no
to rate and post a comment
Dina Alnahdi
re: re: re: Brilliance 16
April 10, 2019 04:58
The 7 segment led display shows number "3"
to rate and post a comment
raj venugopal
re: re: re: re: Brilliance 16
April 11, 2019 11:52
sir you send the error detail and your mobile no my mail id rajvenugopal@gmail.com
to rate and post a comment
Dina Alnahdi
re: re: re: re: Brilliance 16
April 11, 2019 11:52
I forget to tell that this problem happened when I switch to service key of the gantry and then use the up & down switches at the bottom of the table. May be some thing happened to one of the limit switches. My question is how many limit switches do I have? Because I recognize 6, and not sure if there is any more?
to rate and post a comment