Biosound megas problem

(21)

(1)

Anh Tran

Biosound megas problem

February 05, 2010 10:53

When I change the TCG ( level 3 max or level 6,7 min) , there are white vertical lines in the middle ot the screen so I can't make the scan. If i change the TCG, sometime these lines become very small but till have.
Anybody know that? Please help me resolve this problem. thanks

Log inor Register

to rate and post a comment

(39)
(4)
(14)

Thomas Vella

Re: Biosound megas problem

February 06, 2010 11:32

could be dead elements in the probe check with another probe to see if the problem goes away
hydracomp inc
hydracomp@aol.com

Log inor Register

to rate and post a comment

(21)

(1)

Anh Tran

Re: Biosound megas problem

February 07, 2010 10:02

Thanks for your suggest, I did check the probe already, and sure that the probe is OK, anything else I need check?
Thanks in advance

Log inor Register

to rate and post a comment

(39)
(4)
(14)

Thomas Vella

Re: Biosound megas problem

February 07, 2010 01:43

reloading the software would be my next choice before opening up the system,if you have to try reseating the boards they shift always and can cause problems
tom

Log inor Register

to rate and post a comment