Over 500 Total Lots Up For Auction at One Location - NJ 06/15

ATL Ultramark 400C Problem

Tags: Ultrasound, ATL

Posted by an Anonymous User

ATL Ultramark 400C Problem

March 10, 2016 01:05

Hello, at the end of the booting process the system displays this error message: Checksum of C:/PUBCOLOR/system/probe/atl/prob_p37.pd0 is not correct

Does anyone has an idea? Thanks

Log inor Register

to rate and post a comment

(119)
(57)
(6)

Jim Carr

re: ATL Ultramark 400C Problem

March 11, 2016 11:04

That could be a hard drive or corrupted software problem. Try restoring the backup of the presets. Worst case is that it could be caused by a front end board.

Log inor Register

to rate and post a comment

(12)

Alberto Osuna

re: re: ATL Ultramark 400C Problem

March 11, 2016 03:56

It is a software problem, I reload the software and everything is OK but I do not have the options (COLOR, 3D, DICOM, etc) when I try to write the option code in the mod pass the system displays that message and do not boot, do you know how to set the options on the new HDD?

Log inor Register

to rate and post a comment

(161)
(20)
(4)

Miguel Chavira

re: ATL Ultramark 400C Problem

June 02, 2021 02:10

your software is bad
you need reload the software in a new hdd

Log inor Register

to rate and post a comment