artifacts philips ct

sunil Kumar

artifacts philips ct

December 30, 2007 08:17

i am getting rings in air scans and steaks in patient scan.they are more pronounced in abdomen scans near pelvic region.head scans ,they are very light and almost non existent in upper cuts.can it be bad calibration??

Log inor Register

to rate and post a comment

saju bhaskaran

Re: artifacts philips ct

December 30, 2007 01:42

Dear Sunil
what system is this ?LX ,mx or brilliance?
with regrads
saju

1 person liked this
Log inor Register

to rate and post a comment

(46)

(5)

Charles Psarreas

Re: artifacts philips ct

December 30, 2007 01:43

this sounds like a bad chan
do a scout in the large field
it you see a line artifact , this is your bad chan.

what model is it?