Hello!
Help solve the problem with the crescent-shaped artifact on Aquilion32
On the scan it is presented in the form of a white strip.
Appears after some time of work.
If you turn off the computer and turn it on again, the artifact disappears
Please did you check DAS power supply- Detector Temperature- Air calibration must not be old. Please can you test both reconstruction rack ? Also check if a red led lit on SCRT or GCIF PWB when artifact appears ( this system is MUDAT or SureCom?
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
evg evgen
Aquilion32 Help!
April 14, 2017 05:50
Hello!
Help solve the problem with the crescent-shaped artifact on Aquilion32
On the scan it is presented in the form of a white strip.
Appears after some time of work.
If you turn off the computer and turn it on again, the artifact disappears
to rate and post a comment
QuantaMed
re: Aquilion32 Help!
April 17, 2017 04:43
Please did you check DAS power supply- Detector Temperature- Air calibration must not be old. Please can you test both reconstruction rack ? Also check if a red led lit on SCRT or GCIF PWB when artifact appears ( this system is MUDAT or SureCom?
to rate and post a comment