Error Code --> 6c44e
E_T14_PMC_1_START,
Meaning : PMC1 I/O-board could not be started successfully.
This board is responsible for driving all kinds of
actuators aboard the Drystar printer.
Reason : Board does not respond correctly during its
initialization phase (at start-up).
Cure : a. Check cables and board. Use CAN-analyser to
verify the CAN-communication if judged
necessary.
b. Check power-supply side as well, check the
power-cables leaving for the PMC-board.
c. Check BER-board sitting on top of the CPU-board.
Error 6C44E:
Meaning: PMC1 I/O-board could not be started successfully. This board is responsible for driving all kinds of actuators aboard the Drystar printer.
Reason: Board does not respond correctly during its intialization phase (at start-up).
Cure: a) check cables and board. Use CAN-analyser to verify the CAN-communication if judged necessary.
b) Check power supply side as well. check the power cables leaving for the PMC-board.
c) Check BER board sitting on top of the CPU board.
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
Leo Samba
Agfa drystar 3000 ERROR 6C44E
April 04, 2013 02:10
I have ERROR 6C44E CALL SERVICE on agfa drystar 3000. Need help
to rate and post a comment
Konstantin Kuznetsov
re: Agfa drystar 3000 ERROR 6C44E
April 05, 2013 10:05
Error Code --> 6c44e
E_T14_PMC_1_START,
Meaning : PMC1 I/O-board could not be started successfully.
This board is responsible for driving all kinds of
actuators aboard the Drystar printer.
Reason : Board does not respond correctly during its
initialization phase (at start-up).
Cure : a. Check cables and board. Use CAN-analyser to
verify the CAN-communication if judged
necessary.
b. Check power-supply side as well, check the
power-cables leaving for the PMC-board.
c. Check BER-board sitting on top of the CPU-board.
to rate and post a comment
tim cowell
re: Agfa drystar 3000 ERROR 6C44E
April 05, 2013 10:10
Error 6C44E:
Meaning: PMC1 I/O-board could not be started successfully. This board is responsible for driving all kinds of actuators aboard the Drystar printer.
Reason: Board does not respond correctly during its intialization phase (at start-up).
Cure: a) check cables and board. Use CAN-analyser to verify the CAN-communication if judged necessary.
b) Check power supply side as well. check the power cables leaving for the PMC-board.
c) Check BER board sitting on top of the CPU board.
to rate and post a comment
Konstantin Kuznetsov
re: re: Agfa drystar 3000 ERROR 6C44E
April 05, 2013 01:42
Forgive for that I outstripped you, Tim. I give you the way.
Good luck...