Hello Mike,
The next explications is referent your error in the CR85-X...
226C4
E_IBF_TST_SNBELT_POWER
Reason : Input buffer module: Power supply test of stepper node Belt failed.
Meaning : No Power supply on stepper node detected.
- power supply not connected.
- fuse blown.
- power supply defective.
Cure : - check cables and connectors.
- check/replace fuse.
- check/replace power supply.
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Mike Rodemers
AGFA CR85-X
August 20, 2013 04:42
Does anyone know what error code 226C4 means upon power up of CR 85-X. Please advise or if you have service manual with error codes for this unit
to rate and post a comment
Rebel Cook
re: AGFA CR85-X
August 20, 2013 07:10
Power supply test of stepper node belt failed. Check the cables and connectors, fuses, power supply, safety interlocks.
to rate and post a comment
Luis Diaz
re: AGFA CR85-X
August 20, 2013 07:10
Hello Mike,
The next explications is referent your error in the CR85-X...
226C4
E_IBF_TST_SNBELT_POWER
Reason : Input buffer module: Power supply test of stepper node Belt failed.
Meaning : No Power supply on stepper node detected.
- power supply not connected.
- fuse blown.
- power supply defective.
Cure : - check cables and connectors.
- check/replace fuse.
- check/replace power supply.
Good luck!
Bye
to rate and post a comment