Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

AGFA CR 35-X error code

Tags: PACS / DR / CR / Digital, AGFA

(76)
(25)
(7)

Massiera Gilles

AGFA CR 35-X error code

April 23, 2014 09:17

Hi there,

We have a CR 35-X AGFA who works good but sometimes when we turn it on, we have the 20E0D error code after auto test.
Does someone know this error or has the service manual please ?

Thanks in advance

Log inor Register

to rate and post a comment

(58)
(18)

Rebel Cook

re: AGFA CR 35-X error code

April 23, 2014 10:07

The PMT is out of range, basically seeing to much light. If you have the side panel off the cr during startup testing it can cause this error. If the cover is on then check the wiring to the PMT. I have had two machines with this error and both had bad PMTs. If this is a newer unit that has the acrylic light guide then you have to replace the light guide and PMT as one assembly.

1 person liked this
Log inor Register

to rate and post a comment

(76)
(25)
(7)

Massiera Gilles

re: re: AGFA CR 35-X error code

April 24, 2014 08:03

Thanks a lot, I will check all that, hoping we won't have to change PMT...
Thanks again

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: AGFA CR 35-X error code

April 24, 2014 01:13

Error Code --> 20e0d
E_SCN_FN_TIV_P_RANGE,
Meaning : Pixel values at photomultiplier test mode are out of range.
Reason : Photomultiplier or cable defective or disconnected.
(Possibly also: ht_pm_test wrong calibrated)
Cure : Avoid too much background light, keep door closed at start.
Check cable from scan control board to photomultiplier box.
If problem persists, replace photomultiplier box.

Log inor Register

to rate and post a comment

(76)
(25)
(7)

Massiera Gilles

re: re: AGFA CR 35-X error code

April 24, 2014 08:25

Thanks a lot, I will check all that, hoping we won't have to change PMT...
Thanks again

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Source for Sony Medical Displays, Printers & More!

Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.

stats

Événements de Vente aux Enchères à Venir

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)