Dimitris Afantitis

ACCURUS ERROR

April 01, 2019 09:12

907 and 903 error in accurrus vitrectomy system

please advice

Log inor Register

to rate and post a comment

(1)

(14)

Dr.Phaco

re: ACCURUS ERROR

December 20, 2019 09:49

Sorry for the delay in responding. These are voltages that are out of tolerance. You will need a new power supply.

The Accurus may use one-of-two power supplies. Power Supply II part number 202-1460-501 is assembled by Alcon, and its theory of operation is included here.

Power Supply III part number 202-1617-001 is a vendor-supplied product, and all information for this power supply is provided on the drawing included in section 6 of this manual.