Hello team technicians, please can you help me in FUJI AC-3, i have error 03c2 , i have service manual, but i don't find problem, maybe SZ1 but where is it ??? thanks your help in advance.
Regards
rdus0001@yahoo.fr

Log inor Register

to rate and post a comment

(9)
(4)
(5)

Indarpersad Singh

re: AC-3, error 03c2

January 24, 2014 08:19

Look at page 416 of your manual. Parts List for sub scanning Unit for location of SZ1. It is #23