JOSE HERRERA
May 22, 2025 10:52
Hello colleagues I need your help I have a problem with the detector of a mammography machine brand Siemens model Mammomat MG Inspiration Software ...
Armando Zarate
October 22, 2024 10:18
Hi I have a Mammomat inspiration that is giving me rectangles in the Magnification images even in the calibration I only get images like that Does any...
Michael Fisher
July 02, 2013 06:01
I recently replaced the tube on a GE 2000 D mammo unit and it calibrated accordingly However when I performed the Flat Field it failed MTF and ROI I...
Joe bello
October 22, 2013 09:08
I have error 094004 image aborted by the IDC I have calibrated the x-ray tube AOP conversion factor measurement mas non-linearity detector gain check...
Michael Buck
October 02, 2016 02:53
Have a Hologic Multicare Platinum Need to replace Microprocessor BD and want to back up cal data Where to hookup Laptop and what commands to use to ...
Svilen Popov
August 20, 2019 10:31
Hello I have problems with a GE Senographe 800T showing the following error E12 Checksum Error 071081 I changed the ram-back up battery but had this e...
islam hassaan
July 21, 2019 01:13
Dear all I need service manual for GMI Lambada s mamography with calibration manual I check my machine today with various doses with different M...
Greg Togtema
February 25, 2018 11:10
Trying to do gain calibrations on a senograph essentials machine Specifically I am trying the stereo calibrations The manual says to insert the ster...
Lenny Place
May 29, 2018 03:55
I am looking for the following service documents for the Selenia MNL SEL CALIBRATION AND SW MAINTENANCE MAN-00746 Calibration and Software Mainte...
JERONIMO ROLDAN
October 25, 2017 09:15
Hi Dear colleagues I try to calibrate the Bias voltage in the Senographe 2000D but when measuring with the multimeter between C and G and press the se...
Julian Aguilera
May 17, 2017 10:28
Hi everyone I have a 800T senographe with error E06 068025 floating calcul Address 220EBC When I turn on the mammo passes the test well but after 5 ...
Diego Salgado
April 26, 2017 10:33
Someone with the service manual for mammo LORAD Elite i need fix the error 61 appears on the screen can help me this is my email phsmeridayahoocom...
Yojin Ramones
May 26, 2025 04:22
Dear all I have a white line in chest side with 18x24 padle and mag padle It si looking the 1 or 2 mm of blade collimator with 24x30 is ok Some could...
Rainner Lopez
June 11, 2024 11:26
Hello everyone im looking for the service software and Service Manuals for this unit it would be of great help if someone can send it to me Rainner...
agfa drystar problem
james garza
April 18, 2023 03:53
Hi colleagues A very interesting problem at the Agfa 5302 when doctors send a command for printthe medical image film is printed very well and no pro...