AKMAL TURAEV
February 26, 2017 02:39
Hi all i got a problem with installation of OS to CT LightSpeed QXI OCTANE2 console Can you help me with DVD-ROM model exactly which model is used in...
TH Khan
October 17, 2014 05:58
Dear seniors As my Positioner NVRAM battery expired i changed it with new one the i flash it with my CPU firmware Zidane Now my my terminal conne...
Manuel Esquer
September 26, 2014 04:42
After install software LFC M4 in GE CT HISPEED QXi H2 system reboots then pass Autostart and then shutdown immediately Found in log file a prob...
Roberto Angel Vaca Valverde
September 26, 2014 10:13
We have a Innova 2000 Adavantax error the system shows a error code at the moment of the exposure 42C4 tube anode not rotating try again On the DL Ap...
Leonti Pan
September 23, 2014 08:17
OS will not boot processing unit Hiris A12 from x-ray equipment Prestige SI Leave a message UMB upper limit segment address used when working normally...
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
EFREN FRANCO
September 12, 2014 09:57
I have an error M 050 system initialisation problem in a C-arm Philips Endura at start up If anyone can help me with suggestions or technical informat...
Johnny Barreto
September 08, 2014 06:10
Im getting big trouble servicing this machine The system is a GE Prodigy Advance with sw version 1140 It has a CSBC board and a DIB board I didnt have...
David Gerarduzzi
February 19, 2019 10:12
Dear all we are having problems with a GE innova 4100 cath lab At the start up the system shows three problems at the room monitor - xray no...
Peter Russell
August 08, 2014 05:56
Hi I have a problem with a merge frame grabber on start up the Linux red had program stops with the message AN ERROR OCCURED DURING THE FILE SYSTEM CH...
Calvin Lavy
March 09, 2016 03:46
hii bought a ge voluson 730 expertand it Boots up without any issues and detects transducers but does not display an image or doppler The transducer e...
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Bill Masters
July 17, 2014 01:24
GE Collimator 2292592- Under Table Getting the following errors log- Reads as this Error Code 600524- Under Table coll error height limit switch ...
Posted by an Anonymous July 01, 2014 08:40
hello i have voluson 730 BT4 appear line on image on probe image this lines some times come and some times gone i hope if you can help me and tel...
MARIO RAUL GONZALEZ
June 20, 2014 07:51
problem during daily cal only starts self test and locating the QA block after that only a couple of scans and says no block found in the error log 22...
Senad Sejdinovic
June 10, 2014 01:52
Hi im working on Philips Gemini TF 64 system And im having some problems with CIRS system it contains 10 separate CIRS servers Basicly 3 of CIRS st...
Serial number error in Siemens Acuson S2000
Tuy Do Duy
October 10, 2016 10:58
Dear all I have trouble with my ultrasound System can not load software and I reload it But when after I reload software system have trouble it displ...