Hi guys, i have a problem with Sterilclave 18B/24 - autoclave manufactured by COMINOX Italy.
The cycle start normally and work good till 25-30min, after that it show an error "hermetic alarm"
and the clave not continue working, just suspend here.
Can anyone help to resolve this problem. I can upload pictures or videos.

Log inor Register

to rate and post a comment

(246)
(33)

CRAIG WAMBURA

re: Sterilizer Sterilclave 18B/24

April 14, 2019 04:07

You should be having a 'INSUFFICIENT STEAM ALARM' indicating that the steam is leaking

But the chamber seal is hermetic meaning its airtight .... so you should be having an issue with your chamber seal.

Check your door gaskets...
clean internal disk on the door....remove any product residue....
Check the tilt of the sterilizer...(review installation manual)
If alarm still on....replace seal..

Contact for further instructions: craigwambura@outlook.com +256782702243 (WhatsApp)

Log inor Register

to rate and post a comment

(30)
(4)

Michel Bessette

re: Sterilizer Sterilclave 18B/24

April 14, 2019 04:08

The most common fault with this style of autoclave is the door gasket not sealing.

C. Michel Bessette, RRT (Canada #72-659)
Servicios Tecnicos CMB7
16 Crescent
Kanata, Ontario. K2M 1L1
serviciostecnicoscmb7@gmail.com
Mobile: 0016138185321

Log inor Register

to rate and post a comment

(8)
(1)

AMADI

re: Sterilizer Sterilclave 18B/24

April 14, 2019 04:08

first confirm that the thermostatic valve is working well

Log inor Register

to rate and post a comment

(8)
(1)

AMADI

re: Sterilizer Sterilclave 18B/24

April 14, 2019 04:08



check the thermostatic valve it seems to be loosing pressure

Log inor Register

to rate and post a comment

(5)

Borislav Ivanov

re: re: Sterilizer Sterilclave 18B/24

May 08, 2019 10:22

Hello thanks for your help, can you tell me where is locate this thermostatic valve ,its EV1 to EV5 , and how can i check this work or no?