Hi
For C-Arm Siemens Siremobil 2000, what is
err 0021 and how to fix
Also I need the Service manual and trouble shooting.
Thanks in advance.

Log inor Register

to rate and post a comment

(208)
(27)
(27)

Wayne Horsman

re: Siremobil 2000

September 11, 2017 02:58

0021 I.I. power supply error
Cause : Message from I.I. power supply. The power supply for the image intensifier (high voltage) is out-of-tolerance.
Active: Constantly
Possible trouble / help : I.I. power supply
. Interruption in indication lead from I.I. to board D6; operating voltage missing (also triggers error 0020).

email me if you need a replacement poor supply.
wayne@columbiaimaging.com

Log inor Register

to rate and post a comment

(3)

Alaa Abd elLateef

re: re: Siremobil 2000

September 11, 2017 04:51

Thanks alot for your kind reply...
I already fixed the problem and now every thing is OK..
Many thanks again
Alaa Mohammed

Log inor Register

to rate and post a comment

(208)
(27)
(27)

Wayne Horsman

re: re: re: Siremobil 2000

September 12, 2017 09:20

for my benefit next time this comes up, what did you find?

Log inor Register

to rate and post a comment

(3)

Alaa Abd elLateef

re: re: re: re: Siremobil 2000

September 13, 2017 09:06

I got the Service manual few hours before going for repaire;so I followed the error led indicators in D11;it was OK.. the green led.
So I checked the led in D120; the green led wasn't light after the self test...I checked all connections...the socket which feeds the board was loosen..so I fixed it and every thing became OK
Thanks

Log inor Register

to rate and post a comment

Servicing GE Nuclear Medicine equipment with OEM trained engineers

We offer full service contracts, PM contracts, rapid response, time and material, camera relocation. Nuclear medicine equipment service provider since 1975. Click or call now for more information 800 96 NUMED

 

 

 

L'Access et l'utilisation de cet emplacement est sujet aux modalités et aux conditions du notre de nos MENTIONS LEGALES & DONNEES PERSONELLES
Propriété de et classe des propriétaires DOTmedà .com, inc. Copyright ©2001-2017 DOTmed.com, Inc.
TOUS DROITS RÉSERVÉS