I have a Logiq A-200 ultrasound using MST board 2211942-7. The box is not booting and previous posts said probably the board is bad.

I found a 211942-8 board - can I put this board in without any issues? With the next version / revision (8) be able to just be put in and I do not have to upgrade anything else on the box?

Thanks

Log inor Register

to rate and post a comment

(349)
(103)
(2)

Marcelo Antonio Correa

re: Logiq A-200 MST 2211942-7 board replacement

February 24, 2020 09:23

Hi, Kevin!

Prior to replace the board, replace the MST battery.

Over the same board you will find dip switches, place number one to on postition, and turn the L200 on.

Built-in SW installer will load the EEPROM.

After the boot, turn off L200, and return the dip switch to off postition.

Clock will probably shows date of 1998.

If you're not familiar on how to enter service mode and set date and time, call me inbox, please.

Additionally, the board you found will probably work, too.

Hope these help, take care, God Bless.