dismiss

Clean Sweep Live Auction on Wed. February 27th. Click to view the full inventory

SEARCH
Endroit courant :
>
>
>
> Logiq 200 Ultrasound Screen Brightness Issue

Logiq 200 Ultrasound Screen Brightness Issue

Tags: Ultrasound, GE

(1)
(1)

Laurence Boss

Logiq 200 Ultrasound Screen Brightness Issue

January 31, 2018 09:27

I'm currently trying to fix a LOGIQ 200 Ultrasound in a provincial hospital in Cambodia. The CRT monitor displays at a very high brightness making it unusable and the brightness dial has no effect on the screen brightness.
Any help on this problem would be greatly appreciated.

1 person liked this
Log inor Register

to rate and post a comment

(120)
(8)

Ramon Portillo

re: Logiq 200 Ultrasound Screen Brightness Issue

February 13, 2018 08:08

Hi Laurence, your problem is a resistor that is broken, I mean open. If you write to me I will email the service manual.

Your problem is common in CRT monitors, so check the resistors where the grid or Bias... The problem will be in a voltage divider circuit. Anything you may contact to me at: ramon.portillo.m@gmail.com

Log inor Register

to rate and post a comment

GE Healthcare ultrasound probe solutions to help protect your investment

GE Healthcare helps reduce the downtime of ultrasounds with online access to probe and accessory purchasing, probe-failure diagnoses, fast repairs and replacements, probe care and handling tips, and reliable service options.

 

advertisement

 