GE OPTIMA 660 DETAIL CALIBRATION

February 12, 2018 05:14

Dear colleagues:
I had a conversation with Chilen technologist that said me the process to calibrate the Optima 660 with a NON CENTERED phantom is incorrect.

Service manual point that I must rise table 40mm from laser center to calibrate SMALL and other 40mm to LARGE. Of course, images obtained are not centered and look cut upside.
Am I interpreting manual erroneously?

Best Regards
José

Log inor Register

to rate and post a comment

(19)
(7)

Dragos Mosneagu

re: GE OPTIMA 660 DETAIL CALIBRATION

February 14, 2018 09:29

Follow the manual :)) you are right !