GE AMX 4 Plus X-Ray

February 24, 2020 09:29

I have GE AMX 4 X-Ray. Problem is all Initial Test 1, Test 2, Test 3, Test 5, failed. Please help to to resolve the problem.

Log inor Register

to rate and post a comment

(97)
(14)

Donald Bradbury

re: GE AMX 4 Plus X-Ray

February 25, 2020 10:24

First off, what is the condition of your battery set? Can you charge the batteries? If the set is over 3-4 years old they are suspect. Then you can attempt a recalibration. Make sure you understand the process before starting as once you begin, the proceedure must be successfully completed to avoid further problems. I recommend using invasive kv measument as opposed to non-invasive to avoid errors at the lower kv cal points that can give you problems. You should expect to attempt repeating the cal process a few times till you get the hang of it. Good luck.

Log inor Register

to rate and post a comment

(42)
(7)

CharlieWheel

re: GE AMX 4 Plus X-Ray

February 25, 2020 10:24

Maybe your AMX is in short circuit, It can be the Batteries array or high voltage card. Initially I would review these two points.

Log inor Register

to rate and post a comment

(61)
(17)

tim cowell

re: GE AMX 4 Plus X-Ray

February 26, 2020 09:31

TEST-01 IS EPROM CHECK SUM
TEST-02 IS RAM BATTERY
TEST-03 AUXILIARY RAM
TEST-05 WATCHDOG TIMER
REPLACE CPU AND RE-CALIBRATE

1 person liked this
Log inor Register

to rate and post a comment