Fuji Fcr Prima question for Error 270

October 12, 2016 10:12

Hello. I have the following situation: I have Error 270, but only Warning(11270). I change to show it only in Error Log. Do I need to do something? To clean the lighting guide, or calibrate S-value ?

Best Regards...

Log inor Register

to rate and post a comment

(280)
(50)

Gabriel Cool

re: Fuji Fcr Prima question for Error 270

October 13, 2016 11:02

No. Its just warning and its not important. You can disable this massage in RU settings

Log inor Register

to rate and post a comment

(101)
(7)

Svilen Popov

re: re: Fuji Fcr Prima question for Error 270

October 18, 2016 06:39

Okay, I disable it. But what is this warning? Why it appear in RU ?

Log inor Register

to rate and post a comment

(1)

Karan Nair

re: re: Fuji Fcr Prima question for Error 270

May 13, 2019 10:30

Hi i have a problem regarding error 252 in prima t reader where it is an intermittent issue,so kindly suggest me on which parts do i need to replace or what solution do i need to offer to the customer

Log inor Register

to rate and post a comment

(101)
(7)

Svilen Popov

re: re: re: Fuji Fcr Prima question for Error 270

May 15, 2019 10:27

Error 10252: [During bootup or routine processing]
The image reading error occurred.

Check the error log for other errors. If any error occurs, analyze it.
- Check the CPU board fuses. Replace the PSU27A power supply,
the PSU27B board and CPU board if the fuse blowout recurs
immediately after replacement of the fuses.
- Reseat the connectors connecting with the scanning optics unit.
- Replace the scanning optics unit.
- Replace the SED board.
- Replace the cables connecting from the SED board to the CPU
board.
- Replace the CPU board.

Log inor Register

to rate and post a comment

ProCare - Let us keep your MRI system up and Running with Confidence!

ProCare ensures that the medical equipment you rely upon performs at an optimum level! Our engineers take pride in fast response time leading to less down time! Let us keep you Running With Confidence! Call 855.367.4330 to speak direct with an engineer.

 

 

advertisement

 

advertisement

 

advertisement
advertisement
L'Access et l'utilisation de cet emplacement est sujet aux modalités et aux conditions du notre de nos MENTIONS LEGALES & DONNEES PERSONELLES
Propriété de et classe des propriétaires DOTmedà .com, inc. Copyright ©2001-2019 DOTmed.com, Inc.
TOUS DROITS RÉSERVÉS