February 13, 2018 08:08

Hello.
Can someone help me with an explanation of the error SERVICE 24724? Tnx.

Log inor Register

to rate and post a comment

(79)
(5)

Ronald Swapp

re: CR 85-X

February 13, 2018 04:40

Error is:
E-IOC_ERACTL_OPENRELAIS
Reason: IOBUS control module: Erasure control board safety test:
Safety Relais does not open again.
Meaning: Safety relais on Erasure control board for interrupt power supply
to erasure lamps does not open.
Hardware Problem.
Cure: Replace Erasure control Board.

Send me an e-mail and I'll send you the service manual for the 85 and IMOS Show Error Viewer.
rswapp@comcast.net
Authorized AGFA Field Service Tech.

1 person liked this
Log inor Register

to rate and post a comment

(79)
(5)

Ronald Swapp

re: CR 85-X

February 13, 2018 04:40

My guess is you lost the Erasure Control Board, GS102 number: CM+9515564204.
Or you may have a problem with the Current Sense Board, GS104. part number: CM+9515564401.
Go for the GS102 First, I believe there is a micro relay that has a contact that is welded closed from over current.

Log inor Register

to rate and post a comment

OiHealthcare - ISO 9001 & 13485 Certified - Call 1-888-278-9933

Oxford Instruments Healthcare, is a leading ISO 9001 & 13485-certified organization, that specializes in providing quality refurbished CT & MRI systems, service, mobile units & parts at prices you can afford. Call 888-278-9933 Email DOTmedinfo@oxinst.com

 

 

 

L'Access et l'utilisation de cet emplacement est sujet aux modalités et aux conditions du notre de nos MENTIONS LEGALES & DONNEES PERSONELLES
Propriété de et classe des propriétaires DOTmedà .com, inc. Copyright ©2001-2018 DOTmed.com, Inc.
TOUS DROITS RÉSERVÉS