Salvador III Gaviola
Siemens Multix Pro, Err 0009
March 26, 2019 10:18
Hi to all,
Any guidancein rectifying Err 0009 appears in X-ray, Siemens Multix Pro during booting the machine.
Thanks in advance.
to rate and post a comment
Bill Masters
re: Siemens Multix Pro, Err 0009
March 27, 2019 10:03
B.Masters@bmx-ray.com
Err009
D510 Error
Description:
⢠A D510 inverter error has occurred.
Possible Causes of Error:
⢠W520 cable defective or contact problems
⢠Power Supply defective
⢠D510 defective
Corrective Measures:
⢠Check low voltages (Wiring Diagram).
⢠Check the ribbon cable between D500 X10 and D510 X10 (W520); if needed, replace
⢠Replace the inverter
to rate and post a comment
Germni
re: Siemens Multix Pro, Err 0009
March 27, 2019 10:05
Repeating watchdog reset ==> Replace D50.
to rate and post a comment