I need your help,machine displays error8 on prep.
Fuse F101 found blown, replaced and blew again on prep.Connector K20 checked found properly connected.
Can anyone with more experience help me? Thank you.

Log inor Register

to rate and post a comment

Bill Masters

re: Siemens Mobilette II

September 27, 2018 06:05

Remove U101 from unit and measure it like a common fuse.
Warning: Discharge the capacitor bank before touching U101 !!!

If you find the fuse blown, first of all order a new one (I would order more than one in your place...).
Before replacing the fuse with a new one, I would recommend a thorough check of all the capacitor blocks for problems (e.g. blown capacitor fuses).

If the new fuse blows again (especially in high kV values) then you should suspect your x-ray tube...