I have problems with X-Ray system on Philips Brilliance CT 6.
Modules HSS, R-Host, M-Host, G-Host and power supply of +/-15V and +5V for R-Host & Vicor +24V have been tested and confirmed OK.
Sometimes I got theses errors:
1)Filament Calibration ERROR
2)ALE-KV_E-Stop_140kV with Tube Disconnected on No load test
3)APM ARC with 140kV with Tube Disconnected on No load test
4)140kV Error on Air-Calibration
All Manual Exposure tests with 90kV from 50 to 500mA with LARGE Focal point pass OK.
All test with SMALL Focal point failed.
SCREENSHOTS from tests can be seen here(incl. Grid Settings):
Hello Eugene,
unfortunately I don't work with Philips CT scanners, so I can't help. By the way I just wanted You to know that You need to modify the dropbox link You provided: it should end with the number "1", not "0", in order to permit the download of the ZIP file You have prepared with the PNG pictures inside of it. I just tested it: empty page with finishing "0", downloaded the ZIP file with finishing "1".
I just wanted to tell You, so that hopefully someone else will be able to see the pictures and help You. ;->
My friend, your Tube is arcing. Is probably dead. You can try a reloading, but is not guarantee that it will work. Is a long shoot, but is cheap comparing with a brand new CT Tube....
Good Luck!
re: re: re: Need help with X-ray on Philips Brilliance CT
January 28, 2017 03:56
S_RMC_SB_FAULT ( when Gantry Init )
S_GENARATOR_MONITOR_GAF_ERROR ( when Tube Conditioning )
S_RHOST_HVG_HSS_GENATOR_NON_ARC_FAULT & S_MASTER_SEQUENCER_EXIT_STATE_ERR ( when Filament Calibration )
I have changed APM & HSS (there is NO more ARCS)
The final steps should be:
45)Run the required calibrations according to the applicable Brilliance System Calibration manual.
46)Test the system to verify that it is operating properly.
I have checked XRT grid voltages - they are OK.
The machine does NOT pass the Filament Calibration now.
On R-HOST after the errors I see CODE:15 and one constant yellow light ( no more blinking lights as before )
link for the pictures with errors:
https://www.dropbox.com/sh/bcj80w2po5jf2g8/AADTx8UoaAFxVoecbyd2-Qvga?dl=1
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Eugene Genoff
Need help with X-ray on Philips Brilliance CT
January 17, 2017 10:25
I have problems with X-Ray system on Philips Brilliance CT 6.
Modules HSS, R-Host, M-Host, G-Host and power supply of +/-15V and +5V for R-Host & Vicor +24V have been tested and confirmed OK.
Sometimes I got theses errors:
1)Filament Calibration ERROR
2)ALE-KV_E-Stop_140kV with Tube Disconnected on No load test
3)APM ARC with 140kV with Tube Disconnected on No load test
4)140kV Error on Air-Calibration
All Manual Exposure tests with 90kV from 50 to 500mA with LARGE Focal point pass OK.
All test with SMALL Focal point failed.
SCREENSHOTS from tests can be seen here(incl. Grid Settings):
https://www.dropbox.com/sh/s4vcniw8do2cj5k/AAD_-Xb6E3tj7T_R8LDR_Wj9a?dl=0
Any suggestions how should I proceed ?
genapic@yahoo.com
to rate and post a comment
Ivano Murat
re: Need help with X-ray on Philips Brilliance CT
January 18, 2017 07:43
Hello Eugene,
unfortunately I don't work with Philips CT scanners, so I can't help. By the way I just wanted You to know that You need to modify the dropbox link You provided: it should end with the number "1", not "0", in order to permit the download of the ZIP file You have prepared with the PNG pictures inside of it. I just tested it: empty page with finishing "0", downloaded the ZIP file with finishing "1".
I just wanted to tell You, so that hopefully someone else will be able to see the pictures and help You. ;->
Best regards,
Ivano Murat
to rate and post a comment
Serkan OZTAS
re: Need help with X-ray on Philips Brilliance CT
January 18, 2017 03:58
I can supply all service and training docs if you need. serkan.oztas@rtsistanbul.com
to rate and post a comment
Gustavo Volpi
re: Need help with X-ray on Philips Brilliance CT
January 23, 2017 07:57
My friend, your Tube is arcing. Is probably dead. You can try a reloading, but is not guarantee that it will work. Is a long shoot, but is cheap comparing with a brand new CT Tube....
Good Luck!
to rate and post a comment
Eugene Genoff
re: Need help with X-ray on Philips Brilliance CT
January 25, 2017 04:02
Thank you all for your king support.
There is some progresses.
Does somebody knows what
MCU 65 Error means?
I could find only Part Number : 4535-665-00231 for the MCU but no Error codes in my manuals?
to rate and post a comment
Gustavo Volpi
re: re: Need help with X-ray on Philips Brilliance CT
January 25, 2017 05:49
Check the Error Log, you'll find the complete Error Code. Share with us and we can help you...
to rate and post a comment
Eugene Genoff
re: re: re: Need help with X-ray on Philips Brilliance CT
January 28, 2017 03:56
S_RMC_SB_FAULT ( when Gantry Init )
S_GENARATOR_MONITOR_GAF_ERROR ( when Tube Conditioning )
S_RHOST_HVG_HSS_GENATOR_NON_ARC_FAULT & S_MASTER_SEQUENCER_EXIT_STATE_ERR ( when Filament Calibration )
I have changed APM & HSS (there is NO more ARCS)
The final steps should be:
45)Run the required calibrations according to the applicable Brilliance System Calibration manual.
46)Test the system to verify that it is operating properly.
I have checked XRT grid voltages - they are OK.
The machine does NOT pass the Filament Calibration now.
On R-HOST after the errors I see CODE:15 and one constant yellow light ( no more blinking lights as before )
link for the pictures with errors:
https://www.dropbox.com/sh/bcj80w2po5jf2g8/AADTx8UoaAFxVoecbyd2-Qvga?dl=1
to rate and post a comment
nilo panlilio,jr
re: re: re: re: Need help with X-ray on Philips Brilliance CT
March 03, 2017 09:34
replaced your X-ray Tube.
to rate and post a comment
Eldad Tor
re: Need help with X-ray on Philips Brilliance CT
March 19, 2017 05:03
Hi we can help you.
dean@mtcmed.com
to rate and post a comment