centrifuge 5804r showing error 3

August 20, 2014 08:25

My Eppendorf centrifuge 5804R showing no rotor and error 3 when start is pressed. The rotor dont move at all but can hear the compressor sound when press start. Please help.

Log inor Register

to rate and post a comment

(3)

(9)

Jacob Reinhart

re: centrifuge 5804r showing error 3

August 21, 2014 12:51

Here is what the service manual says about error 3

Error 3
no rotor
No rotor pulses during run or
acceleration.
Missing rotor, measuring system for
rotational speed defective, drive
defective.
– Rotor not fastened.
– No pulses after start. Device brakes
for 15 s.
– Pulse failure during run. Free deceleration
requires 12 min!
Error remains active after a power
supply interruption.
– Fasten rotor.
Keep device switched on and start
again after > 12 min.
– Service program 8.
– Check/replace motor cap with speed
sensor, check connections.
– Replace "Control/FC" PCB,
check connector X10 on PCB
"Control".
– Check run of motor, replace motor if
necessary.

Log inor Register

to rate and post a comment

(2)

Long Nguyen Kim

re: re: centrifuge 5804r showing error 3

September 07, 2016 12:13

My Eppendorf centrifuge 5415R showing Error 3 and can not control anything when I turned the switched on.
Please help me fix this error! Thanks!

Log inor Register

to rate and post a comment

(3)

(9)

Jacob Reinhart

re: centrifuge 5804r showing error 3

August 21, 2014 12:51

Error 3

Cause

no rotor
No rotor pulses during run or
acceleration.
Missing rotor, measuring system for
rotational speed defective, drive
defective.
– Rotor not fastened.
– No pulses after start. Device brakes
for 15 s.
– Pulse failure during run. Free deceleration
requires 12 min!
Error remains active after a power
supply interruption.

Remedy

– Fasten rotor.
Keep device switched on and start
again after > 12 min.
– Service program 8.
– Check/replace motor cap with speed
sensor, check connections.
– Replace "Control/FC" PCB,
check connector X10 on PCB
"Control".
– Check run of motor, replace motor if
necessary.

Let me know if we can help.

Log inor Register

to rate and post a comment

(2)

Long Nguyen Kim

re: centrifuge 5804r showing error 3

September 07, 2016 12:14

My email: longnk@atilab.com.vn

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Medical Equipment Superstore Surplus Auctions July 14th

Medical Equipment Superstore closeout auctions take place on DOTmed.com. Bid today on 276 lots of medical equipment. Items can be inspected in person or via video call. Contact Rick at +1 (215) 669-3913 or info@medicalequipmentsuperstore.com

stats

Événements de Vente aux Enchères à Venir

Medical Equipment Superstore
TIMED
Jul 14, 2026 (501 lots)