BV Pulsera Mobile C-arm doesn´t get ready message M326
June 10, 2026 12:39
Hello, dear members of forum.
I have a Philips Pulsera Mobile C-arm system. Recently, I have been facing up some error codes. It is M326 and M050. After checking it, I found the fan in ESU unit stuck. It was already replaced, but equipment is not working yet.
The whole battery set was recently replaced, but after a 24-hours charging cicle, battery voltage is just +339 Vdc. I know that it should be +360 Vdc. Could someone help me to solve out this issue? I´ll really appreciate any help. Thanks in advance.
re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 11, 2026 10:48
Thanks a lot for replying, Leon.
No.
I have already checked all connections between all batteries, besides that, battery voltage was checked one by one, measured voltage is +12.89 Vdc to +13.01 Vdc on each when the equipmente is OFF. Equipment was working ok for one week with all neww batteries.
Have any other suggestion?
Regards
re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 11, 2026 04:09
Please Open the charger and set the charging voltage to 365V with the Potentiometer. Check the internal resistance of every lead battery, they should be all from one producing charge with identical internal resistance. Best regards Holger
re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 11, 2026 06:34
Hi, Holger.
Thanks a lot for replying.
Let me tell you, the maker of ESU unit is Gilardoni.
I tried to adjust voltage from +339V to +360V with R215, but not success, the voltage value didn´t change. Measuring points were J11, J12 battery connections, equipment turned ON.
How can I check the internal resistance of every battery?
Best regards!
re: re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 12, 2026 11:49
hi Anyfe, you need introduce the new battery in the system configuration to recognize the new battery, but you need to do this with a service software and password.
re: re: re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 12, 2026 05:00
Thanks a lot for answering, Ricardo.
If the system don´t recognize a new battery set, could that be the cause of M326 error or not battery voltage increasing?
Regards
re: re: re: re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 12, 2026 10:33
yes because the system recognize the old battery, if you don't change the new battery you have the error. I don't can hep you with the sw.. I don't have.
I had the same problem.
re: re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 15, 2026 09:40
R215 is correct to adjust the charge voltage to 360-365V (system on, without x-ray), in use the battery-voltage should be over 330V, if not, Error 326 or 397 come, then one ore more batterys have too much internal resistant. We check internal resistant with DUOYI DY223, each lead battery must have about 0,030 Ohms (30mOhms)
re: re: re: re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 19, 2026 02:00
hi AnyFE
Table 16 SUCO - X-ray generator control pulse board - LED indicators
3.11.4.5.3 Fuses
None.
3.11.4.5.4 Jumper settings
LED Color Label Description
H1 A/B Red Error Indication of the test results during power ON.
H2 A/B Green Run
H3 A Green +5 V ON = +5 V DC present
H3 B Green +3.3 V ON = +3.3 V DC present
H4 A Green +15 V ON = +15 V DC present
H4 B Green -15 V ON = -15 V DC present
H5 A Yellow “Safety” ON = Hand/foot switch command active
H5 B Yellow “En X sw” ON = X-ray software command present
H6 A Yellow “HFSACM” ON = Hand/foot switch safety command active
H6 B Yellow “RD PR X” ON = Ready prepare X-ray command present
H7 A Yellow “CTRL_X” ON = Control X-ray command present
H7 B Yellow “HVCOM” ON = High voltage command present
H8 A Yellow “RADPR” ON = radiation present
H8 B Yellow “X ACT.” ON = X-ray actual command present.
I hope it works for you.
re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 12, 2026 11:50
Good morning again. Then there must be an issue with a bundle from the charger board, going to wherever it goes. Check those two. We are doing that now as well with a new battery set from Gilardoni, and tracing every bundle and fuse in the process
re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 12, 2026 05:37
Thank you very much for replying, Leon.
Actually, I found the fan in charger unit damaged. It was already replaced but system keeps M326 error yet. Every bundle and battery was checked, as connections too, fuses are ok...everything on ESU unit seems to be ok.
Regards.
re: re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 12, 2026 10:33
AnyFe, are you sure that the error is M326??
326--BV-scope. x ray generator not available, check cable between generator control board and SE unit. generator control board has failed replace board. SHA4.
re: re: re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 15, 2026 09:40
Hello, Ricardo.
Yes, I´m sure. M326 error. All connections to and from ESU unit were already checked. How can I check SHA4 board?
Thanks a lot for replying.
Regards
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
AnyFE
BV Pulsera Mobile C-arm doesn´t get ready message M326
June 10, 2026 12:39
Hello, dear members of forum.
I have a Philips Pulsera Mobile C-arm system. Recently, I have been facing up some error codes. It is M326 and M050. After checking it, I found the fan in ESU unit stuck. It was already replaced, but equipment is not working yet.
The whole battery set was recently replaced, but after a 24-hours charging cicle, battery voltage is just +339 Vdc. I know that it should be +360 Vdc. Could someone help me to solve out this issue? I´ll really appreciate any help. Thanks in advance.
Regards!
to rate and post a comment
Leon Gugel
re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 10, 2026 02:37
you might have a disconnected cable within the battery wiring set. Recheck all the connections there, and from the charger board as well
to rate and post a comment
AnyFE
re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 11, 2026 10:48
Thanks a lot for replying, Leon.
No.
I have already checked all connections between all batteries, besides that, battery voltage was checked one by one, measured voltage is +12.89 Vdc to +13.01 Vdc on each when the equipmente is OFF. Equipment was working ok for one week with all neww batteries.
Have any other suggestion?
Regards
to rate and post a comment
Holger Fritze
re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 11, 2026 04:09
Please Open the charger and set the charging voltage to 365V with the Potentiometer. Check the internal resistance of every lead battery, they should be all from one producing charge with identical internal resistance. Best regards Holger
to rate and post a comment
AnyFE
re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 11, 2026 06:34
Hi, Holger.
Thanks a lot for replying.
Let me tell you, the maker of ESU unit is Gilardoni.
I tried to adjust voltage from +339V to +360V with R215, but not success, the voltage value didn´t change. Measuring points were J11, J12 battery connections, equipment turned ON.
How can I check the internal resistance of every battery?
Best regards!
to rate and post a comment
Ricardo Rivera
re: re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 12, 2026 11:49
hi Anyfe, you need introduce the new battery in the system configuration to recognize the new battery, but you need to do this with a service software and password.
to rate and post a comment
AnyFE
re: re: re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 12, 2026 05:00
Thanks a lot for answering, Ricardo.
If the system don´t recognize a new battery set, could that be the cause of M326 error or not battery voltage increasing?
Regards
to rate and post a comment
Ricardo Rivera
re: re: re: re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 12, 2026 10:33
yes because the system recognize the old battery, if you don't change the new battery you have the error. I don't can hep you with the sw.. I don't have.
I had the same problem.
to rate and post a comment
Holger Fritze
re: re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 15, 2026 09:40
R215 is correct to adjust the charge voltage to 360-365V (system on, without x-ray), in use the battery-voltage should be over 330V, if not, Error 326 or 397 come, then one ore more batterys have too much internal resistant. We check internal resistant with DUOYI DY223, each lead battery must have about 0,030 Ohms (30mOhms)
to rate and post a comment
AnyFE
re: re: re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 19, 2026 10:57
THANKS A LOT FOR RESPONDING, HOLGER.
I WILL GET AN INTERNAL RESISTANCE CHECKER TO CHECK EACH BATTERY. WHAT IF ANY BATTERY IS OVER 30mOHMS?
to rate and post a comment
Ricardo Rivera
re: re: re: re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 19, 2026 02:00
hi AnyFE
Table 16 SUCO - X-ray generator control pulse board - LED indicators
3.11.4.5.3 Fuses
None.
3.11.4.5.4 Jumper settings
LED Color Label Description
H1 A/B Red Error Indication of the test results during power ON.
H2 A/B Green Run
H3 A Green +5 V ON = +5 V DC present
H3 B Green +3.3 V ON = +3.3 V DC present
H4 A Green +15 V ON = +15 V DC present
H4 B Green -15 V ON = -15 V DC present
H5 A Yellow “Safety” ON = Hand/foot switch command active
H5 B Yellow “En X sw” ON = X-ray software command present
H6 A Yellow “HFSACM” ON = Hand/foot switch safety command active
H6 B Yellow “RD PR X” ON = Ready prepare X-ray command present
H7 A Yellow “CTRL_X” ON = Control X-ray command present
H7 B Yellow “HVCOM” ON = High voltage command present
H8 A Yellow “RADPR” ON = radiation present
H8 B Yellow “X ACT.” ON = X-ray actual command present.
I hope it works for you.
to rate and post a comment
Leon Gugel
re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 12, 2026 11:50
Good morning again. Then there must be an issue with a bundle from the charger board, going to wherever it goes. Check those two. We are doing that now as well with a new battery set from Gilardoni, and tracing every bundle and fuse in the process
to rate and post a comment
AnyFE
re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 12, 2026 05:37
Thank you very much for replying, Leon.
Actually, I found the fan in charger unit damaged. It was already replaced but system keeps M326 error yet. Every bundle and battery was checked, as connections too, fuses are ok...everything on ESU unit seems to be ok.
Regards.
to rate and post a comment
Ricardo Rivera
re: re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 12, 2026 10:33
AnyFe, are you sure that the error is M326??
326--BV-scope. x ray generator not available, check cable between generator control board and SE unit. generator control board has failed replace board. SHA4.
to rate and post a comment
AnyFE
re: re: re: re: re: re: BV Pulsera Mobile C-arm doesn´t get ready message M326
June 15, 2026 09:40
Hello, Ricardo.
Yes, I´m sure. M326 error. All connections to and from ESU unit were already checked. How can I check SHA4 board?
Thanks a lot for replying.
Regards
to rate and post a comment