URGENT!!! ICONOS R200 ERROR 315/ 048
November 08, 2014 09:39
Hello dear colleagues;
I am looking for help to fix a failing x-ray system. My whole system is like this:
POLYDOROS SX 65/80, AXIOM ICONOS R200, FLUOROSPOT COMPACT, XCU.
When I turn the equipment on, the operator console shows "E 315 AP ID 048 SEVERE ERROR CALL SERVICE". The cassette transport transverse motor(M2) does not move neither the cassette clamping (cassette tensioning) motor (M3). I checked voltage at XP2=1.1 volts respect to XP8 (0V). It must be +24V. I checked on D2 pcb the LEDs B10 (WDT error signal or general error has occurrred), B12 (CE_B enable signal not present -octal driver, motor controller-) and they are lit. Besides that, B15 is not lit, indicating that + 24V is off.
I tried those movements on service mode but nothing happens.
DonĀ“t now where +24V voltage come from. Is there any fuse to check for that voltage? I will thank a lot any help from you. I also checked voltage at XP3=5.01V, XP4=5.0V,that means that K2, K1 are OK.
Best regards.
My e-mail: vicmaal@yahoo.com.mx
to rate and post a comment
Lenny Place
re: URGENT!!! ICONOS R200 ERROR 315/ 048
November 10, 2014 04:34
check F6 and f7 in the power supply 2 section of the M1 in the table electronics.
I will send you an email with diagrams
to rate and post a comment
(10)
Vic MayAl
re: re: URGENT!!! ICONOS R200 ERROR 315/ 048
November 14, 2014 03:44
Hi Lenny,
Thanks a lot for answering, sorry for my delay to replying back. Your support helped me to fix the machine, it is working now; I appreciate that a lot.
Best regards.
to rate and post a comment
Lenny Place
re: re: re: URGENT!!! ICONOS R200 ERROR 315/ 048
November 16, 2014 11:13
I am glad you were able to repair it. Take care