The images show ring artifact, I already did air calibration and the problem persists. I tried to make channel correction, but it shows the following message:
... calculating channel correction values failed, possibly caused by a defective channel...
What are the steps in the software to find and disable the defective channel?
Software version: Somaris/5 VA47C
I appreciate the help you can give me.
I believe he already made that conclusion Eric . I think hes not sure how to troubleshoot the specific detector board. Do you have any tips? I'm no Siemens expert.
Altima will inspire the trust, loyalty and commitment of our customers through positive relationships and collaboration, top quality equipment services, and a clearly superior value proposition. Call Today (469) 644-7063
(4)
Rommel Saavedra
Somatom Volume Zoom Ring atifact
August 09, 2019 10:02
Hello
The images show ring artifact, I already did air calibration and the problem persists. I tried to make channel correction, but it shows the following message:
... calculating channel correction values failed, possibly caused by a defective channel...
What are the steps in the software to find and disable the defective channel?
Software version: Somaris/5 VA47C
I appreciate the help you can give me.
to rate and post a comment
(4)
Eric Hauk
re: Somatom Volume Zoom Ring atifact
August 11, 2019 02:25
Likely bad DAS board, you need to troubleshoot further and figure out which one
to rate and post a comment
Aaron Anderson
re: re: Somatom Volume Zoom Ring atifact
August 18, 2019 11:23
I believe he already made that conclusion Eric . I think hes not sure how to troubleshoot the specific detector board. Do you have any tips? I'm no Siemens expert.